LCR meter and ESR for small capacitors

Discussion in 'General Electronics Chat' started by Elerion, Jan 7, 2018.

  1. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    Hi.
    I've bought my first LCR meter, capable of testing at 100 Hz to 100 kHz.
    It can measure ESR in Cs mode (capacitance, series measurement).
    For electrolytics, it seems to work fine.

    But, if test a good 100 nF ceramic capacitor, it measures hundreds of ohms resistance, which it obviously must be wrong.
    For a good 1 uF ceramic, it tells me 16 ohms (again, wrong, it should be an order of magnitude lower).
    The D factor is so low for all of them. Just the ESR is abnormally high.

    Also, for the some working caps, I sometimes see a lower D factor at some frequency, but a higher ESR.
    This also happends for different electrolytics. Some have a better D factor, but worst ESR. Does it make sense?
    Could it be the meter?
     
  2. Nigel Goodwin

    Nigel Goodwin Super Moderator Most Helpful Member

    Joined:
    Nov 17, 2003
    Messages:
    39,417
    Likes:
    668
    Location:
    Derbyshire, UK
    ESR is for electrolytics, don't worry about it for other types of caps.

    High ESR is by FAR the most common failure mode for electrolytics, and the most common fault in electronics generally.
     
    • Informative Informative x 1
  3. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    Yes, I know. But I wanted to test if the meter is working fine, and in the way, learn a bit more about how they work.

    The discrepancy between D factor and ESR does concern me.
    Isn't a bad ESR capacitor suppose to have a high D factor?
    Why could a higher ESR capcitor read better D factor on a meter?
    There's a relation between them, as far as I know from theory.
    I heard some people replacing electrolytics based only on the D factor, not the ESR.
     
  4. dave

    Dave New Member

    Joined:
    Jan 12, 1997
    Messages:
    -
    Likes:
    0


     
  5. Nigel Goodwin

    Nigel Goodwin Super Moderator Most Helpful Member

    Joined:
    Nov 17, 2003
    Messages:
    39,417
    Likes:
    668
    Location:
    Derbyshire, UK

    I'd never even heard of the term until you mentioned it, and I've never seen an ESR meter that displays it.

    ESR meters are extremely common in the service trade, and I've used them thousands of times locating faulty electrolytics - the other option is to use a scope, assuming the unit powers up for scope readings.
     
  6. kubeek

    kubeek Well-Known Member

    Joined:
    Mar 11, 2006
    Messages:
    1,544
    Likes:
    192
    Location:
    Prague, Czechia (not Chechnya)
    ONLINE
    Try measuring say a 1ohm smd resistor, it should have very little capacitance or inductance, but should show exactly 1ohm.
     
  7. The Electrician

    The Electrician Active Member

    Joined:
    Jan 13, 2009
    Messages:
    558
    Likes:
    71
    What make and model is your new meter?

    You didn't say at what frequency you were making the measurements. What style is your 100 nF caramic? Is it an older disc type, or a more modern surface mount? Try measuring your 100 nF ceramic cap at 100 Hz, 1 kHz, 10 kHz and 100 kHz. What do you get?
     
  8. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    It is a handheld Tenma, capable of 100kHz and ESR measurements.

    I tried all frequencies, from 100 Hz up to 100 kHz, but focussing on 1 kHz.
    100 nF older disc type, yes (leaded). Not multilayer.
     
  9. Nigel Goodwin

    Nigel Goodwin Super Moderator Most Helpful Member

    Joined:
    Nov 17, 2003
    Messages:
    39,417
    Likes:
    668
    Location:
    Derbyshire, UK
    As I said above, you're wasting your time - ESR testing is for electrolytics, not small capacitors.
     
  10. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    Sure. I just answered the questions.
    The main point of all this was testing if the meter was operating correctly.

    Remember it is a LCR meter, not just an ESR meter.
     
  11. Nigel Goodwin

    Nigel Goodwin Super Moderator Most Helpful Member

    Joined:
    Nov 17, 2003
    Messages:
    39,417
    Likes:
    668
    Location:
    Derbyshire, UK
    I do, but it doesn't mean that ESR is relevant for non-electrolytic capacitors.
     
  12. The Electrician

    The Electrician Active Member

    Joined:
    Jan 13, 2009
    Messages:
    558
    Likes:
    71
    What values did you get at the various frequencies?
     
  13. The Electrician

    The Electrician Active Member

    Joined:
    Jan 13, 2009
    Messages:
    558
    Likes:
    71
    You should expect the ESR for an older disc type ceramic cap to be high at low frequencies. Here is an image showing the ESR and impedance magnitude at frequencies from 100 Hz to 5 MHz of a 100 nF ceramic disc cap. The impedance is shown in green and ESR in yellow. Note that the scales are logarithmic.

    The ESR at 100 Hz is 385Ω, at 1 kHz it's 38Ω, at 10 kHz it's 5 ohms and at 100 kHz it's 1.4Ω

    Be aware that the ESR of capacitors of various types typically vary with frequency, sometimes by orders of magnitude. There's nothing wrong with your meter.

    DiscESR.jpg
     
    • Thanks Thanks x 1
  14. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    Thank you! That was exactly what I meant.

    Several hundreds ohm resistance is very high. So, does this mean that ceramic capacitors are unsuitable for low frequency filtering?
     
  15. Nigel Goodwin

    Nigel Goodwin Super Moderator Most Helpful Member

    Joined:
    Nov 17, 2003
    Messages:
    39,417
    Likes:
    668
    Location:
    Derbyshire, UK
    Is this why you've been trying to check ESR of small capacitors?.

    For decoupling you basically stick a high value capacitor across the supply for low frequencies (which won't be any good for high ones), and a low value capacitor across the supply for high frequencies (which won't be any good for low frequencies), the two together cover pretty well all you need.

    No need for tests or measurements, it's simply standard design practice.
     
  16. kubeek

    kubeek Well-Known Member

    Joined:
    Mar 11, 2006
    Messages:
    1,544
    Likes:
    192
    Location:
    Prague, Czechia (not Chechnya)
    ONLINE
    Most esr meters don't discern between esr and capacitive reactance. That is why the esr number looks horrible at low frequency, even though that is the reactance and esr could be much less than an ohm. Will try with two lrc bridges on monday to be sure :)
     
  17. The Electrician

    The Electrician Active Member

    Joined:
    Jan 13, 2009
    Messages:
    558
    Likes:
    71
    It's not the reactance that Elerion's meter is measuring at low frequencies; it is in fact the ESR.

    Elerion's meter isn't specifically just an ESR meter; it's an LCR meter. He didn't give us the model number, but I'm pretty sure it's this one: http://www.newark.com/tenma/72-10465/lcr-meter/dp/26W7021

    The fact that it can measure at frequencies of 100 Hz, 1 kHz, 100 kHz and so forth differentiates it from so-called ESR meters which almost always measure only at 100 kHz. It's an LCR meter, not just an ESR meter.

    The reason Elerion got a high value for ESR at low frequencies is because those old leaded disc capacitors actually have a high ESR at low frequencies. I showed this with the impedance analyzer sweep in post #12. You should get the same result with a bridge. His meter can discern between ESR and capacitive reactance.

    This Tenma meter is one of a breed of meter that various manufacturers have begun selling since the nifty Cyrustek chip became available: http://www.cyrustek.com.tw/product-2-6.htm

    This makes it easy and inexpensive to manufacture a handheld LCR meter covering the standard frequencies of 100 Hz, 120 Hz, 1 kHz, 10 kHz and 100 kHz. Besides the Tenma meter, the DE-5000 meter is a prime example of such a meter that can be bought for less than $100: https://www.ebay.com/itm/DER-EE-DE-...179223&hash=item25ed243ceb:g:JpcAAOSw~QRaLNG0

    These meters based on the Cyrustek chip can distinguish between reactance and ESR at low frequency, medium frequency and high (100 kHz) frequency. They also can measure ESR at 100 kHz, and they can measure capacitance and inductance. They do a surprisingly good job given their low cost.
     
  18. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    It is.
    The meter seems a good alternative to very expensive units.
    But measuring high inductances (several Henry and up) is tricky. The reading value does not stabilize. It very slowly drifts for so much time, I end up taking an estimation. Sometimes the "Lp" value slowly goes down, while the "Ls" value slowly goes up.
    Is this normal?

    No. Just testing the LCR meter.
    I bought the meter to test power and audio output transformers and inductors, and, as I have no esr meter, to also detect faulty electrolytics.
    The DCR measurement for small resistance is better than my true-rms DMM (I just miss that it could make four point test measurements).
     
  19. The Electrician

    The Electrician Active Member

    Joined:
    Jan 13, 2009
    Messages:
    558
    Likes:
    71
    This can happen when you try to measure the inductance of a winding which has an iron core, such as a power transformer winding, or a large crossover inductance with an iron (or ferrite) core.

    The inductance of iron cored windings varies with the excitation level. Since your meter uses a fixed excitation level you can't do the experiment where you can see the different measured inductance as you change the excitation, but it does exist. When your meter suddenly applies a sine wave to the inductance, there is a transient component that decays with time and causes the slow drift in the measurement.

    Try this: hook up your meter to an inductance with iron core and wait for the reading to stop drifting. Then disconnect the meter and reconnect with the leads reversed. You should see a different reading which drifts a lot for a while, then should settle to approximately the same reading you got before reversing the leads.

    There's not really anything you can do to avoid this phenomenon. You just have to realize that measurements of iron cored inductances are very approximate, and will change with time and with the magnetic state of the core.

    With regard to that last bit, try this: measure the inductance of a winding on a small power transformer. Place a decent size magnet in contact with the core momentarily then re-measure the inductance. It will probably be quite different.
     
    Last edited: Feb 12, 2018 at 1:16 AM
    • Informative Informative x 1
  20. Elerion

    Elerion Member

    Joined:
    Dec 1, 2015
    Messages:
    115
    Likes:
    2
    Yes, that's the case. Chokes, power and audio output transformers.
    It also depend on the measurement mode; if measuring series or parallel, which is something I'm still learning how to use properly.

    Oveall the meter is good value for the money, and it is very usefull to measure parasitic inductance and capacitance without having to setup a test circuit, signal generator and oscilloscope.
     