Using an AC voltmeter solves some problems, too, by averaging out a non-sinusoidal notched signal. Using a scope as the measurement tool, where do you measure? Peak-to-peak? Some kind of assumed average? What?
The usual procedure is to measure the RMS value of the residual and compare it to the RMS value of the input waveform. The HP distortion analyzers use an RMS converter IC. Most modern digital scopes can calculate the RMS value of a waveform.
After all, if this is such a wonderful, easy, accurate and low THD level method of measuring THD, then why don't Tektronix, Agilent/HP and all the others use it rather than using some version of a notch filter and measuring the output as they do? Ding, ding, ding, ding ... dang, there goes that meter of mine again.
The author referenced in post #11 certainly didn't call the method "easy"; what he said was:
"Adjustment of the variable components is difficult and takes a lot of time and patience."
That's why the commercial distortion analyzer manufacturers don't use that method.