Just tried again with 3V peak-to-peak 50Hz triangular wave.
With no DC offset, in AC coupling mode, oscilloscope measures Vrms = 879mV and the APPA 207 true RMS multimeter reads 0.8882V. When a 3V DC offset is added, the reading is 885mV on oscilloscope and 0.8932V on the Appa 207. The reading on the APPA 207 remains in agreement with the oscilloscope as the frequency increases to around 250kHz, after that the reading will decrease and eventually reaches 0. The oscilloscope of course is still able to measure Vrms until the frequency reaches its bandwidth limit.
So I guess the true RMS capability on the Appa 207 is quite good.
Interestingly, for both my Rigol DS1052E and HP 54600B oscilloscope, V(avg) of both square wave, sine wave, and triangular wave is measured to be approximate 20-30mV. I think that's due to the ADC errors and otherwise should have been zero. Is that expected? Your previous post says that the average value of a 3.906v peak to peak square wave is 1.953v (50% of the peak to peak).
Also if the oscilloscope is in DC coupling mode, Vrms will read incorrectly on both oscilloscopes and show 4.625V for a 3V peak-to-peak triangular waveform with around 3V DC offset. So in this aspect, both oscilloscopes are worse than the Appa 207, which knows how to exclude the DC offset and shows proper RMS value. Or am I missing something?
With no DC offset, in AC coupling mode, oscilloscope measures Vrms = 879mV and the APPA 207 true RMS multimeter reads 0.8882V. When a 3V DC offset is added, the reading is 885mV on oscilloscope and 0.8932V on the Appa 207. The reading on the APPA 207 remains in agreement with the oscilloscope as the frequency increases to around 250kHz, after that the reading will decrease and eventually reaches 0. The oscilloscope of course is still able to measure Vrms until the frequency reaches its bandwidth limit.
So I guess the true RMS capability on the Appa 207 is quite good.
Interestingly, for both my Rigol DS1052E and HP 54600B oscilloscope, V(avg) of both square wave, sine wave, and triangular wave is measured to be approximate 20-30mV. I think that's due to the ADC errors and otherwise should have been zero. Is that expected? Your previous post says that the average value of a 3.906v peak to peak square wave is 1.953v (50% of the peak to peak).
Also if the oscilloscope is in DC coupling mode, Vrms will read incorrectly on both oscilloscopes and show 4.625V for a 3V peak-to-peak triangular waveform with around 3V DC offset. So in this aspect, both oscilloscopes are worse than the Appa 207, which knows how to exclude the DC offset and shows proper RMS value. Or am I missing something?