Set the scope up for single-channel operation, e.g., Channel 1. For channel 1, make sure you have the scope input DC coupled, the variable volts/div in the "calibrated" detent and the attenuator set for 2V/DIV. Position the trace to the bottom horizontal graticule line. The timebase should be set for about 1ms/DIV with the sweep control set for "automatic" so you have a bright trace without any input signal. Using a X1 probe or simple test leads, connnect the Channel 1 BNC to the battery, positive to the center contact. The trace should deflect upward about 4.5 major divisions.
Connecting this channel to the calibrator output should provide a square wave.
Dean