hi,
Looked at your drawing, where did you connect the probe.?

Also would a minimum pH of 2 thru to 14 be acceptable.?
 
I went back to the first post to see what this was for. I found this.

I am developing pH meter should be working from 0- 14 pH. For this i got a pH sensor from
 
Last edited:
hi,
Look at this option, just shows the basic circuit.

The -Vref should be derived from a voltage reference and be adjustable.

Note: for the MCP607 the recommended -Vsig is -0.3V [not -0.413V], so there is a minimum pH signal limit.
If you used a CA3140 opa this would allow a -0.5V input.
 

Attachments

  • pH01.gif
    24.2 KB · Views: 228
I went back to the first post to see what this was for. I found this.

I also went thru the posts and found this

I do have such solutions capsules of 4,7,10 pH. But what should be the actual calibration steps.
Do i need to change the gain value as time passes or some other method??
 
Looked at your drawing, where did you connect the probe.?

The probes would be connected at the V points. The refernce electrode at V1 pin and the measuring electrode at the V2 point.
 
My understanding is that the input span is -413mV to +413mV, and the output span needs to be 100mV to 1100mV. Below is a circuit that should do this. Op amps need to have CMOS inputs.
 

Attachments

  • pH sensor amplifier.PNG
    31.3 KB · Views: 272
Last edited:
hi Ron,

Are you saying that you want to shift the -413mV thru +413mV input range to, 0V thru +826mV or 0V thru +5V.???

yes i that to do that. Need help for the schematic

I guess we will have to see what the OP finally decides upon.

BTW:
For ref only, you may have this already, if so please ignore.