Rigol DM3068 vs Fluke 87V disparity

Status
Not open for further replies.

Mosaic

Well-Known Member
Hi all:
I am reading a low duty cycle pulsed V (<1%) into a 12V batt and the Fluke 87V (both of them that I have, 1 brand new) reads it about 1 to 2% higher than the Rigol DM3068 (brand new).
Reading a steady battery DC source shows them all matched to 3 decimal places....so something is going on with the averaging/RMS readouts.

Any suggestions on resolving this?

Update:
On the DM3068 by selecting the 200V or 1000V range..the readings match the Flukes, on the DM3068 20V range (which happens to be the auto select in this case) ....they don't.
 
Last edited:
It would appear to be related to the crest. As the battery charged and the int res. dropped so did the crest and the 20V scale is now just .03V off the 200V scale that agrees with the Fluke8&V at hi res.
I guess i'll just stay with the 200V scale to be confident of what's going on.
 
It would appear to be related to the crest. As the battery charged and the int res. dropped so did the crest and the 20V scale is now just .03V off the 200V scale that agrees with the Fluke8&V at hi res.
I guess i'll just stay with the 200V scale to be confident of what's going on.

To be fair it's not something you should really attempt to be measuring, at least not with any expected degree of accuracy.
 