| Hey, I just picked up an old teleqipment S51B Oscilloscope for free. I was just wondering how I can make sure it's calibrated accurately. It works and powers on. I've feed it a simple square wave and it seems to be working ok. | |
| There is a variety of simple tests you can do. Set it to DC and connect a say a 1.5 Volt battery to the input. The line on the screen should go up or down by 1.5V depending on which direction the battery is connected. Repeat with other batteries - ie. 3V, 4.5V, 6V, etc. Connect say 6V AC from a transformer to the input. 6V is 16.97 V peak to peak (ie. 6 * 2 sqrt 2). Your AC is 60 Hz, so you can check the sweep rate calibration by measuring the period of the signal. 60 Hz has a period of 1000/60 = 16.67 millisec. If you have an oscillator and frequiency meter, you can check the sweep calibration at other rates. eg. 1000 Hz has a period of 1 ms. Hope this helps, Len | |
| You will have to compensate your probe as well. The scope should have a little pin or something on it that you can connect the probe to, that pin outputs a square wave. There should be a little slot to adjust the compensation on the probe with a screwdriver. Turn it until the square waves are just that: square. If its over-compensated, you will get very sharp peaks, undercompensated just hte opposite. Quote:
__________________ I'm no electronics god, i just talk too much. | ||
| | (permalink) | ||
Len | |||
Also, unless you can get the workshop manual, I wouldn't bother touching any adjustments, from what I remember they are quite obscure in those Telequipment scopes? - with adjustments affecting each other. For 99% of use there's no need for it to be accurately calibrated anyway!. | ||
| | (permalink) | ||
| Thanks for your tips. Quote:
I've found the Oscilloscope in the Dumpster! From what I see it works perfectly. I don't have the lab manual unforunately. Quote:
thanks | |||
| | (permalink) | ||
So your scope might be more than accurate enough for your needs anyway, considering the actual visible reading accuracy from the screen, if it's within 5% or 0.5% you wouldn't be able to tell the difference. I've been using scopes professionally for 34 years, we've NEVER had a scope calibrated, nor had any desire or need to. Everything in live has tolerences, and things you do should bear that in mind. As already suggested, connect it to a battery and measure it's voltage, compare it to your multi-meter, and see if it's fairly similar, it should be, and that's accurate enough for a scope. Bear in mind, how do you know your meter is accurate?, do you have it professionally calibrated and certified every year? (at great expense!). Digital meters have made people rather complacent, your meter might read 1.765V - but it's only going to be 1% or 2% accurate (even when new!), so you can throw the last digit away, and be very dubious about the next one as well!. | |||
__________________ Experience is directly proportional to the value of the equipment ruined. | ||
| Nigel, The X-Axis line is Slightly Rotated at an angle, is there a setting to get it level and center I tore the cover off, and on the bottom board theres six Variable resistors? They are labeled RV89 - Link RV92 - BLKG RV88 - IPS RV6 - SET Y GAIN RV72 - TIME/CM RV48 - Trig SENS | |
| None of those sound to me like they will help your trace problem. Newer scopes often have an adjustment labelled "trace rotate", some even have it on the front panel. That's the one I would be looking for. JB
__________________ Nothing is impossible for the man who doesn't have to do it himself - Weiler's Law | |
| Don't see anything inside or out like that | |
| Perhaps you could loosen the screws holding the CRT and rotate it by hand. Be sure to turn the power off and discharge the EHT first. Len | |